I have a problem in analysis of laboratory data, when I have to take the logarithm of a physical quantity. I know that logarithm is a dimensionless quantity. I have to logarithmize the physical quantity $F=I/V^2$, where $I$ is the electric current and $V$ is voltage. We know that both current and voltage have uncertainty $\mathrm dI$, $\mathrm dV$. In order to find the uncertainty of $\ln(I/V^2)$, I use the formula with partial derivatives and i find that: $$ \mathrm Δ(ln(F))= \sqrt{\frac{\mathrm (dI)^2}{I^2} + 4\frac{\mathrm (dV)^2}{V^2}} $$ If I choose to write voltage and current in volts and amperes, I take a different result in the value of $$ln(F)=ln(\frac{\mathrm I}{V^2})$$than writing them in kilovolts and nanoamperes, for example, but the error has the same value as you can see! So I have a different relative error depending on unit of measurement!
Isn't that weird? What's the problem?
ang_88
  • 31
Your formula for the error is not correct, because it's not dimensionally consistent. Moreover, I could not find support for the formula you state in the link provided in the comments.
If $F=I/V^2$, $$dF=\left|\frac{\partial F}{\partial I}\right| dI+\left|\frac{\partial F}{\partial V}\right| dV=\frac{1}{V^2} dI+2\frac{I}{V^3} dV$$ The absolute values are due to the fact that errors always add. Therefore dividing by $F$: $$\frac{dF}{F}=\frac{dI}{I}+2\frac{dV}{V}$$
Usually, for complicated statistical reasons, it is more appropriate to add errors in quadrature, so that one should actually do: $$dF=\sqrt{\left(\frac{\partial F}{\partial I}dI\right)^2+\left(\frac{\partial F}{\partial V} dV\right)^2}=\sqrt{\left(\frac{1}{V^2} dI\right)^2+\left(2\frac{I}{V^3}dV\right)^2}$$ Dividing by $F$ one gets the result: $$\frac{dF}{F}=\sqrt{\frac{dI^2}{I^2}+4\frac{dV^2}{V^2}}$$ This is the method they use in the link you provided.
Both of the above formulas may be used depending on context. Note that both are dimensionally consistent (i.e. you only add terms with the same dimensions). I don't know if in your analysis you need the absolute error $dF$ or the relative error $dF/F$, that's for you to decide. But both can be obtained from the above. As to the matter of taking logarithms of dimensional quantities, it shouldn't matter as long as you use consistent units. See also the question linked in the comments.
John Donne
  • 5,009
  • 18
  • 31
I would suggest using the functional approach to error propagation in this case. In general, in the majority of cases, it is what I would favour. For your case, the uncertainty in the $\log$ is,
$$\sqrt{\left[ \log \left( \frac{I+\delta I}{V^2}\right)-\log\left( \frac{I}{V^2}\right)\right]^2 + \left[ \log \left( \frac{I+\delta I}{V^2}\right)-\log\left( \frac{I}{(V+\delta V)^2}\right)\right]^2}$$
which simplifies nicely because of logarithms to,
$$ =\sqrt{[\log(I+\delta I) - \log(I)]^2 + 4[\log(V+\delta V) - \log(V)]^2}.$$
Notice there's no problem with dimensions in the uncertainty: we are taking logarithms of dimensionful quantities, but subtracting two logarithms is the same as taking the ratio of the argument, thus producing a dimensionless parameter. Explicitly,
$$=\sqrt{\log^2 \underbrace{\frac{I+\delta I}{I}}_{\mathrm{dimensionless}} + 4\log^2 \underbrace{\frac{V+\delta V}{V}}_{\mathrm{dimensionless}} }$$
For everything you'd need to know about errors for undergraduate experiments (assuming nothing fancy like Markov Chain Monte Carlo), see:
Measurements and their uncertainties: a practical guide to modern error analysis
JamalS
  • 19,600